Introduction.​

Washing laboratory glassware belongs to the mandatory list of skills and abilities of laboratory personnel. High-quality washing implies the achievement of the main result - good cleanliness of chemical glassware for perfect organic synthesis. A number of requirements are put forward for laboratory glassware, since it must always be perfectly clean. It is recommended to wash laboratory glassware immediately after use. The results of the work performed will depend on the cleanliness of the dishes. It's generally easier to clean glassware if you do it right away. When a detergent is used, it's usually one designed for lab glassware, such as Liquinox or Alconox. These detergents are preferable to any dishwashing detergent that might be used on dishes at home. Typically, detergent and tap water are neither required nor desirable. You can rinse the glassware with the proper solvent, then finish up with a couple of rinses with distilled water, followed by final rinses with deionized water.

The best way to clean up laboratory glassware is using some aggressive detergent which will oxydize pollutions or/and remove them easy. There is a list of detergents, synthesis instructions and use manuals.​

Sulfochromic mixture.​

It is a mixture of concentrated sulfuric acid and potassium dichromate; when sulfuric acid acts on dichromate, chromic anhydride CrO3 is formed. Sulfochromic mixture is one of the strongest oxidizing agents. It is widely used in laboratory technology for washing chemical glassware, and also applied in the bleaching process of reversible photography.​
X8M9nbqo0R
QsZOHEYJ6A
A solution of chromic acid in sulfuric acid (also known as a sulfochromic mixture or chromosulfuric acid) is a powerful oxidizing agent, it can be used to clean laboratory glassware, particularly of otherwise insoluble organic residues. Furthermore, the acid leaves trace amounts of paramagnetic chromic ions — Cr(III) — that can interfere with certain applications, such as NMR spectroscopy. This is especially the case for NMR tubes.

You need:
  • Potassium/sodium dichromate (K₂Cr₂O₇/Na₂Cr₂O₇) - 15 g​
  • Concentrated sulfuric (or nitric) acid (H2SO4/HNO3) - 500 ml​
To prepare the chromium mixture, carefully while stirring, potassium dichromate is slowly added to the sulfuric acid. The mixture gets very hot, and the solution turns dark brown. Potassium dichromate is poorly soluble, long stirring with a glass rod is required. The procedure should be carried out in a ceramic glass in a water (cold) bath to remove heat. If sediment remains at the bottom, then don't worry, it should be so.

Application.
Glassware is rinsed with running water, then poured with a chromium mixture and kept for several minutes, if necessary - a couple of days, and then thoroughly washed in running water. On well-degreased glass, water spreads in a thin layer, without collecting into drops. The chromium mixture can be reused until the color changes to greenish. As a result of the ongoing oxidation reactions of organic substances, chromic anhydride is reduced to chromium (III) sulfate, as a result of which the used chromium mixture gradually changes its color to green.

Safety.
Concentrated sulfuric acid is a highly corrosive substance! Pour only acid into water! Chromium mixture is also corrosive; in addition, hexavalent chromium compounds are toxic and carcinogenic. When you are working with a chromium mixture, you should observe safety precautions and use personal protective equipment. Store the mixture in a fume cupboard or in a tightly closed container (not a rubber stopper!). When washing pipettes and various tubes, the mixture should be collected only with a rubber bulb, in no case with the mouth, to avoid severe burns of the mouth and damage to teeth.​

Potassium permanganate solution.​

XgO68mLArc
You need:
  • Potassium permanganate (KMnO4)
  • Oxalic acid/ sodium hydrogen sulfate/ FeSO4/ Mohr's salt

Application.
A good laboratory glassware detergent is a 4% potassium permanganate solution. Potassium permanganate solution is a strong oxidizing agent, especially when heated and acidified with sulfuric acid; it is poured into dishes, which must first be washed with hot water and cleaned with a special brush. Then a small amount of concentrated sulfuric acid is carefully added, which causes heating, quite enough so that all impurities on the walls are quickly oxidized. Sulfuric acid should be taken in such an amount that after adding its solution temperature was about 50-60 °C. Usually, it is enough to add 3-5 ml of concentrated sulfuric acid to 100 ml of potassium permanganate solution. It is necessary to take sulfuric acid and in no case hydrochloric acid, since the latter is oxidized by potassium permanganate with the formation of chlorine. Brown plaque may appear after washing laboratory glassware with a solution of potassium permanganate. It can be removed by rinsing the dishes with a 5% solution of sodium hydrogen sulfate (NaHSO4), solutions of iron (II) sulfate (FeSO4), Mohr's salt or organic acids, preferably oxalic acid. After that, the dishes are washed with water.

The spent acidified potassium permanganate solution is usually discarded and not reused. If a non-acidified solution was used, it can be used several times.

I would recommend you to make big (3-5 l) bath for laboratory glassware, use empty desiccator or another glass or ceramic dish. You can lay dirty glassware into this bath for 2-3 h for oxidizing pollutions on glass walls (previously clean them up by brush and tap water). After, clean this glassware from potassium permanganate solution by tap water and lay for 0.5-1 h into oxalic acid bath to remove manganese oxides (brown plaque). After, repeat procedure with tap water and with distilled water.

Safety.
The same cleaning practices and precautions for handling acidified potassium permanganate solution as described above for the sulfochromic mixture should be followed.​

Alkali alcohol solution.​

The alcohol-hydroxide cleanser is used to clean glass. It is an effective cleaner.

You need:
  • 60 g sodium hydroxide (NaOH)/potassium hydroxide (KOH);​
  • 500 ml ethanol;​
  • 60 ml DI water;​
  • Polypropylene or glass container (600 ml or greater).​
Prepare the sodium hydroxide solution by mixing the crystals into the water. Then add the ethanol. The mixture gets very hot, be careful. You have to stir the solution with a glass rod until dissolution is complete. Be sure to label the container with the title “Ethanol/NaOH 5:1 cleaning solution”.

Application.
Put the laboratory glassware into the bath to soak for 30 minutes. For spotless surfaces, soak for several hours. Rinse in DI water then blow dry. If cleaning solution is clean, and you intend to use it again, store it in an appropriately labeled container. If not, pour the alkaline solution into the appropriately labeled waste container.

Safety
Wear protective gear: eye protection, chemical coat and nitrile gloves.​

To clean glassware, use the following procedures:​

1. Use 2-3 mL solvent to rinse residual organic compounds from the glassware into a waste beaker. The compounds should be highly soluble in the solvent. The default solvent is often acetone as it is inexpensive, relatively nontoxic, and dissolves most organic compounds. Some institutions reuse their acetone ("wash acetone") as the solvation ability is not spent after a few uses.

2. As it will soon become second-nature for most students to use acetone as part of their cleaning ritual, it is worth reminding that the purpose of acetone rinse is to dissolve organic residue in a flask. Not everything dissolves in acetone, for example ionic salts are insoluble in acetone and are more successfully rinsed out with water. After a preliminary rinse, glassware should then be washed with soap and water at the bench.

3. Residual acetone will likely evaporate from the flask, but it is acceptable for small quantities of residual acetone to be washed down the drain. Acetone is a normal biological byproduct of some metabolic processes.

4. If using undiluted detergent from the store, it is best to use small amounts during washing, as they tend to form thick foams that need lots of rinsing. Some institutions instead use dilute soap solutions at their cleaning stations for this reason. For cleaning of glassware, the biodegradable detergent "Alconox" is the industry standard.

This step can be replaced with another one from the list of detergents above. Soap cannot wash any organic substances and one day you will be faced with remains of dirt. The best way is Sulfochromic mixture, which can oxidize almost all organic substances for appropriate exposition time. If you can't find so a lot of sulfuric acid/nitric acid, I would recommend you take Alkali alcohol solution. It is cheap and easy to produce cleaning solution. Potassium permanganate solution is more effectively cope with organic pollutions, but it is quite difficult, and you have to clean up a glassware from manganese oxides by helps of second solution of Oxalic acid, which have to be removed by tap or distillated water.​

Steps in the washing of the laboratory glassware:
5eJwn4udC6

Automatic dishwasher.​

In an underground laboratory, automatic dishwasher can be used as well. But there are a number of problems which can appear with organic or aggressive substances. Plastic membranes, filters, rubbers will crack after some hundred hours of works. Metal surface will rust in acidly environment. To avoid these issues and prolong dishwasher life, you have to rinse glassware by tap water before load it. Your dishwasher will break with a high probability, but if you accept this cost for clean glassware, you can use it. In case of large laboratory load, it helps to safe a lot of time.​
DZwi7gtpsz

Drying Glassware.​

Quick Drying.
If dry glassware is not needed right away, it should be rinsed with distilled water and allowed to dry overnight (in a locker). If dry glassware is promptly needed, glassware can be rinsed with acetone and the residual acetone allowed to evaporate. Rinsing with acetone works well because water is miscible with acetone, so much of the water is removed in the rinse waste. Evaporation of small amounts of residual acetone can be expedited by placing the rinsed glassware in a warm oven for a short amount of time, or by using suction from a tube connected to the water aspirator. Residual acetone should not be evaporated inside a hot oven (>100 °C) as acetone may polymerize and/or ignite under these conditions. It should also not be evaporated using the house compressed air lines, as this is likely to contaminate the glassware with dirt, oil, and moisture from the air compressor.

Oven and Flame Drying.
Glassware that appears "dry" actually contains a thin film of water condensation on its surface. When using reagents that react with water (sometimes violently!), this water layer needs to be removed. To evaporate the water film, glassware can be placed in a 110 °C oven overnight, or at the least for several hours. The water film can also be manually evaporated using a burner or heat gun, a process called "flame drying". Both methods result in extremely hot glassware that must be handled carefully with tongs or thick gloves.​
IGicM1vg5B
a) Removal of vinyl sleeves, b) Clamped, c) First moments with a flame, d) After the flame.​

To flame dry glassware, first remove any vinyl sleeves on an extension clamp (Fig.1 a), as these can melt or catch on fire. Clamp the flask to be dried, including a stir bar if using (Fig.1 b). Apply the burner or heat gun to the glass, and initially fog will be seen as water vaporizing from one part of the glassware condenses elsewhere (Fig.1 c). Continue waving the heat source all over the glassware for several minutes until the fog is completely removed, and the glassware is scorching hot (Fig.1 d). If the glass is only moderately hot, water will condense from the air before you can fully exclude it.

Safety Note: glassware will be extremely hot after flame drying.

Regardless of the manner in which glassware is heated (oven or flame drying), allow the glassware to cool in a water-free environment (in a desiccator, under a stream of inert gas, or with a drying tube, Fig. 2) before obtaining a mass or adding reagents.​
H9IiDCBuTW
Drying Tubes.
A drying tube is used when moderately but not meticulously dry conditions are desired in an apparatus. If meticulously dry conditions are necessary, glassware should be oven or flame dried, then the air displaced with a dry, inert gas.
Drying tubes are pieces of glassware that can be filled with a drying agent (often anhydrous CaCl2 or CaSO4 in the pellet form) and connected to an apparatus either through a thermometer adapter (Fig. 3 b and c) or rubber tubing (Fig.3 d). Air passing through the tube is removed of water when it comes in contact with the drying agent. Since it is important that air can flow through the drying tube, especially so the apparatus is not a closed system, the drying agent should be fresh as used drying agents can sometimes harden into a plug that restricts airflow. Drying tubes can also be filled with basic solids such as Na2CO3 to neutralize acidic gases.​
FRVi8ourcT
a) Drying tube with anhydrous CaCl2, b) Inserting drying tube into a thermometer adapter, drying tube attached to a: c) round bottomed flask, d) distillation apparatus.​

Conclusion and Important facts:​

  • Should clean the glassware as soon as possible.​
  • Regarding delay, put the glassware in water.​
  • As to late cleaning, the residue may not be possible to remove.​
  • New glassware that is slightly alkaline needs to be soaked in acid water (1% HCl or HNO3) for several hours before washing.​
This topic contains the list of most common glassware cleaning techniques. I do not suggest extremely dangerous washing solutions, such as Piranha and so on here because they can cause burns, fire, explosion in low skill chemist arms. Use personal protection in every use of these aggressive detergents and keep your glassware clean, good luck!​
 
Last edited by a moderator:

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
Is there a possibility to clean the alcohol-hydroxide cleanser again to recover at least the solvent?

What do you think of the combination of ethanol and HCL? Would this also be an effective cleaning agent?
I heard in connection with glass frits in filters that it would work well to first suck through hydrochloride acid under vacuum and then a few times ethanol.
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
I think yes, but it's quite inconvenient and take a lot of time, alcohol not so rare and expensive solvent
You will burn your lungs by HCl gas. HCl vaporizing rather fast, I don't recommend this.
Use ethanol at first. Also, don't use alkali to clean up glass filters and ground glass, by reason strong alkali burn their surfaces (slowly).
 
Last edited by a moderator:

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
Ok, I understand the point of view, but I want as little waste as possible. I have no problem neutralizing slightly dirty acids or bases and then disposing of them down the drain, but dumping gallons of solvent is not my cup of tea.
Wouldn't it work to bring the NaOH to Ph6-7 with HCL and then distill the solution? As a result, I would have alcohol that I can reuse and an oily salty shoddy residue which is minimal in volume and weight. I have a way to properly dispose of small amounts of waste (incinerate), so I would just add it to that.
In a pinch I would also have a gas mask :D
I was only concerned about the cleaning effect.
So glass frits can be cleaned in this way very effectively I had to determine, at least the result was a snow-white frit, which was previously brown/red, whether now analytically clean I can not say.
One could prepare a 20% HCL/ethanol solution and rinse glass equipment with it, and neutralize the used rinsing solution with NaOH and distill it again to avoid waste.
Or do I have a mistake in thinking?

When I write HCL I do not mean the pure gas but the 35-37% solution. Sorry if this causes misunderstandings.
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
NaOH alcohol solution? Yes, but it takes a lot of HCl. Also, you will saturate your alcohol with water from HCl aq solution. Forget about it, man.
Use sulfuric acid, it's more effectively. I've never met this way, i don't recommend. You will spend more money for NaOH and HCl soln. Anyway, you can try and write here your result.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I'll definitely try the alcohol NaOH solution, but I'll have to come up with something for the disposal.

I will keep the HCL + alcohol method for glass frits, here you can collect acid and alcohol separately.

Whether this can be transferred to other glass ware remains to be tested.

Sulfuric acid is too precious for me, because it is not normally available.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I have an idea how to recover the ethanol so far.
I now assume that I have 1 liter of ethanol:NaOH solution.
I distill this, as a result I get about 96% ethanol.
Then I salt out most of the water in the separating funnel with potassium carbonate. 2-3 times with approx. 10g potassium carbonate.
Finally I dry with molecular sieve.
With this I should get the ethanol waterproof and clean. Opinions?
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
The best option for glass frits
Laboratory waste disposal
Try and say. But i can't understand why do you want to recover cheep alcohol.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I found the instructions with hydrochloric acid and ethanol at a manufacturer, that works quite well.
Does not the frit then turn purple by using Potassium permanganate? Do I have to rinse with anything?
Yes according to the instructions I do the same, I just do not have combustion or a final storage. I have to see how I get rid of my waste, so I recover everything that is reusable. I don't just dump it anywhere. I can dispose of a small part anonymously and professionally, but only very small amounts and a large part of this is the heavy metal residues.
Sure, it's cheap, but what's wrong with using it more often, if the effort is kept within limits?
Do I make a solvent cleaning day when the containers are full. No problem.
Ethanol would not be an environmental pollution, but in the washing liquid there is also waste that could be disposed of, which should not necessarily be released into the environment.
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
I've used this method. Oxalic acid solution remove everything after washing.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I'll try to get both and then I'll give that a try.

The NaOH/ethanol solution is very good and can be used more often. I hope I will be successful in reprocessing the ethanol so that I can use it again as a washing solution.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
The concentration of the oxalic acid solution is also 5%, or not?

I have the two chemicals here now and can try it out. I am still excited about the alkaline alcohol solution and am already curious how it works.

One thing bothers me a little though, potassium permagnate is very hazardous to water, I don't want to just dump it down the drain. Do you have a simple way to make the potassium permagnate harmless to dispose of it via the drain?
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
correct
Oxalic acid make potassium permanganate to MnO2, which is harmless for wastewater.
 

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I have now tried it, so really convince me it has not, especially because you have to be so incredibly careful with the potassium permanganate, a few small crystals next to it and it turns everything purple. The same with the liquid, you have to remove every drop immediately, otherwise you quickly have a purple lab coat.
If I don't see a big advantage in the potassium permanganat solution, I stick with the alkaline alcohol, the results are impressive, with relatively small use of cheap and easy to obtain materials.

So I fill the potassium permanganat solution into the glass under room temperature and let it stand for several hours, overnight, empty the glass and rinse with the oxalic acid. Or do I also need to heat the solution for better cleaning?
I specifically do not use sulfuric acid.
Approximately how often can I use the solution?
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
1. Put a dirty glassware into a large vessel with potassium permanganate solution
2. Let it stay for ~12 h or so
3. Clean up this glassware by tap water
4. Put this glassware into another large vessel with the oxalic acid for 15-20 min
5. Clean up this glassware by tap water and distilled water.

No heating, no anything else. Very simple. You can use old or cracked desiccator as a vessel.
 
Last edited by a moderator:

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
Yes that's almost how I did it, unfortunately I only have about 1.5 liters of the solution, so I can't really make a bath, at 4% potassium permanganate makes over 40g of the costable good.
But the flask was completely filled and stood overnight, then the solution back into the bottle, rinsed with a little water and you could see a slight brown haze on the glass.
Oxalic acid was added and you could see how the brown disappear.
Oxalic acid back into the bottle and flask rinsed first with tap water, then with distilled water and dried.
Only Unfortunately, the slightly white cloudiness from the flask had not completely disappeared.
Whether it is due to the quality of the glass I can not exclude, of course. Perhaps the cleaning also comes too late.

How often can the solution be used, is there an indicator, by the color I will probably not be able to tell, right?
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
You can put it back in potassium permanganate again.
Use this solution several times. Potassium permanganate solution will be discolored a little bit, and you understand when you need to make new one.
 
Last edited by a moderator:

malayboy

Don't buy from me
Resident
Joined
Apr 15, 2022
Messages
58
Reaction score
35
Points
18
hi sir, I checked out the "Alconox" products turns out they do have a line of products for any laboratory applications, is it ok if I use just all their products of Citranox, Alconox, Liquinox, and Tergazyme, instead of making my own detergent solution like you suggesting sulfochromic mixture, potassium permanganate solution, and alkali alcohol

because in my condition, I am worry about the waste I washing/disposing would corrode the sink underground, then I saw those product of alconox, they claimed those will not rot the sink and free rinsing
 

G.Patton

Expert
Joined
Jul 5, 2021
Messages
2,704
Solutions
3
Reaction score
2,850
Points
113
Deals
1
It wouldn't corrode your sink in case of appropriate manipulations. Also, alcohol/alkaline mixture as well, very well decision. I can't say any exactly about their efficiency. I'm used to preparing solutions by myself for cleaning goals.
 
Last edited by a moderator:

ACAB

Don't buy from me
Resident
Language
🇺🇸
Joined
Mar 27, 2022
Messages
291
Reaction score
168
Points
43
I can highly recommend the alkali/ethanol mix, and also an HCL/ethanol (1:4, my own creation) mix works wonderful, for disposal I just mix the two and add additional fresh water, and otherwise I save those in different containers and use those cleaning solutions several times.
The potassium permanganate solution also works, because there is a slight brown coloration that you get removed with the oxalic acid again, but is compared to the other two a little more complicated.

Whether the purchased solutions achieve a better effect, I can not say, but I see no reason for not using my homemade solutions further.
 